TP200 fault finding
The CMM will not register a probe trigger, but the probe operates normally when the stylus is deflected by hand:
PI 200-3 indicators
Possible causes
Remedy
'STOP' lamp ON
'TP200' lamp ON
'SEATED' LED operates normally
The CMM controller or a Renishaw system has activated STOP signal.
SCR200 overtravel mechanism is deflected.
Check status of Renishaw motorised probe head or other systems.
Clear obstruction and allow overtravel mechanism to reset.
The probe fails to trigger and the probe LEDs glow only dimly when the stylus touches the workpiece, but the probe operates normally when the stylus is deflected by hand:
PI 200-3 indicators
Possible causes
Remedy
'SEATED' LED ON
The trigger speed is too slow.
The stylus is too heavy.
Probe normally to the workpiece surface.
Increase gauging speed.
The probe will not arm or the probe does not stay armed when the RESET button is released. The probe LEDs are always OFF:
PI 200-3 indicators
Possible causes
Remedy
'STD' LED ON
'SEATED' LED OFF
Probe sensor faulty.
Probe wiring open circuit.
Remove probe and test by substitution.
Check wiring from probe to PI 200-3 interface.
The probe will not arm or the probe does not stay armed when the RESET button is released. The probe LEDs are always ON:
PI 200-3 indicators
Possible causes
Remedy
'TP200' LED ON
'SEATED' LED OFF
Probe sensor faulty or damaged by collision.
False (‘air') triggers occur while the CMM is stationary and the probe LEDs flicker:
PI 200-3 indicators
Possible causes
Remedy
'TP200' LED ON
Probe sensor faulty
Probe loose in probe head
Excessive vibration from external source
Excessive vibration from CMM
Correctly tighten probe
Remove cause or isolate CMM
Check CMM air supply
Maintain CMM air bearing system
False (‘air') triggers occur at gauging speed and the probe LEDs flicker:
PI 200-3 indicators
Possible causes
Remedy
'DAMPED' LED is OFF
Stylus is too large or heavy
Use stylus arrangements within recommendations.
False (‘air') triggers occur at traverse speed and the probe LEDs flicker:
PI 200-3 indicators
Possible causes
Remedy
'DAMPED' LED is ON
Traverse speed is too high
Reduce traverse speed
The probe triggers during an SCR200 stylus change:
PI 200-3 indicators
Possible causes
Remedy
The SCR200 is not connected to the PI 200-3
Incorrect SCR200 operating mode
Check SCR200 indicator lamps
Re-connect cable
There is an unexpected loss of accuracy:
PI 200-3 indicators
Possible causes
Remedy
'TP200' LED ON
Stylus ball is damaged or dirty.
Stylus is too large or heavy.
The probe is loose or not correctly assembled.
The kinematic coupling is damaged or dirty.
The gauging speed has been changed.
The trigger threshold has been changed.
Inspect and clean stylus ball, or replace and re-qualify the stylus.
Check the stylus joints. Ensure the module is correctly seated and the probe is tight in the probe head.
Inspect and clean the kinematic coupling.
Re-qualify stylus tips.
Deflection alarm active:
PI 200-3 indicators
Possible causes
Remedy
Audible indicator ON
The stylus is or was deflected for >10 s.
Stylus module was changed manually.
Move the stylus clear of any obstruction and press the RESET button