Troubleshooting
This section on troubleshooting is a guide to problems associated with the installation and integration of the system only. Refer to the 'PH10 PLUS installation and user's guide' (Renishaw part number H-1000-5070) regarding problems associated with normal operation of the system.
The optional HCU2 can also be used to identify system faults. For full details of the use and fault finding capability of the unit see the 'HCU2 hand control unit user's guide' (Renishaw part number H-1000-5361).
Use the table below to identify problems you are experiencing with your system. If you experience problems which you are not able to identify or solve satisfactorily, please contact Renishaw for further advice.
Power on LED not lit
Possible cause
Power loss
Check mains cable connections and integrity.
Check power block is supplying 24 V.
Check CMM emergency stop condition - power may have been removed by CMM.
No head movement in automatic mode
Possible cause
Power loss
See above.
Cable / connection fault
Check connection and integrity of cabling between head and controller.
Communications failure
Check RS232 baud rate.
Incorrect command set selected
Check correct command set is selected.
No head movement in manual mode
Possible cause
Power loss
See above.
No output signal received by the CMM computer
Possible cause
Check connection and integrity of cabling between controller and interface.
Check connection and integrity of cabling between interface and computer.
Multiwire bypass connector not fitted
Fit a multiwire bypass connector with will permit a standard touch-trigger probe signal to reach controller
Multiwired probe in use
Check that the multiwire cable is correctly fitted to the head.
Check that the trigger output to the CMM computer is connected to the multiwired probe interface.
Poor measurement performance
Possible cause
Loose mounting of head
Ensure all mounting screws are tight and mounting to CMM is secure.
Probe damping enabled during measurement
Ensure probe damping is not enabled during measurement moves.