PH10M-iQ PLUS troubleshooting
Overload error
Possible cause
Head lock up holding force exceeded during CMM acceleration
Reduce mass of probe combination.
Reduce acceleration.
Collision with workpiece or CMM
Check head for damage.
Remove obstruction and repeat move.
Obstruct error
Possible cause
Head / stylus configuration obstructed
Probe extension combination too long
Use a shorter / lighter stylus combination.
Internal head fault
Contact your Renishaw representative.
PH10 PLUS head and non-PLUS PHC10-3 or third party controller
Datum error
Possible cause
Cable / connections faulty
Check condition of cabling from head to controller.
Remove obstruction and update head using HCU2 or under CMM control.
Use shorter / lighter stylus combination.
Head damaged
No head movement in automatic mode
Possible cause
Head controller in manual mode
Change to automatic mode.
Power loss
Check mains cable connections and integrity.
Check power block is supplying 24 V.
Check CMM emergency stop condition - power may have been removed by CMM.
Cable / connection fault
Check connection and integrity of cabling between head and controller.
Communications failure
Check RS232 baud rate.
Incorrect command set selected
Check correct command set is selected.
PICS STOP is asserted by other system component
Clear the fault asserting PICS STOP.
No head movement in manual mode
Possible cause
Head controller in automatic mode.
Change to manual mode.
Power loss
No probe output signal received by the CMM computer
Possible cause
Check connection and integrity of cabling between controller and interface.
Check connection and integrity of cabling between interface and computer.
Multiwire bypass connector not fitted
Fit a multiwire bypass connector with will permit a standard touch-trigger probe signal to reach controller
Multiwired probe in use
Check that the multiwire cable is correctly fitted to the head.
Check that the trigger output to the CMM computer is connected to the multiwired probe interface.
No probe signal
Possible cause
Probe incorrectly attached
Remove probe and reattach.
Probe failure
Probe output disabled by CMM
Check probe output.
Poor measurement performance
Possible cause
Loose head mounting
Ensure all mounting screws are tight and that mounting to the CMM is secure
Force applied to head during lock up
Lock and unlock the head
Incorrect lock up position
Reposition head correctly
Position not qualified
Check qualification information
Probe damping enabled during measurement
Ensure probe damping is not enabled during measurement moves
NOTE: Before seeking technical assistance, please make a note of the following information:
  • Your CMM manufacturer and model
  • The probe head model
  • The probe head serial number
  • The head controller serial number
  • The head controller version number
  • The LED sequence on the PHC10-3
  • The stylus configuration fitted
If you experience problems that you are not able to identify or solve, please contact your Renishaw representative for further advice.
