PHC10-2 Troubleshooting
This section on troubleshooting is a guide to problems associated with the installation and integration of the system only. Refer to the PH10 series user’s guide (Renishaw part number H-1000-5070) regarding problems associated with normal operation of the system.
The optional HCU1 can also be used to identify system faults. For full details of the use and fault finding capability of the unit see the HCU1 hand control unit user’s guide (Renishaw part number H-1000-5016).
Use the table below to identify problems you are experiencing with your system. If you experience problems which you are not able to identify or solve satisfactorily, please contact Renishaw for further advice.
Observation
Possible cause
Checks / remedies
POWER ON LED not lit
Power loss
Check mains cable connections and integrity
Check PHC10-2 fuse (see page 6)
Check CMM EMERGENCY STOP condition - power may have been removed by the CMM
No head movement in automatic mode
Power loss
See above
Cable/connection fault
Check connections and integrity of cabling between head and PHC10-2
Communications failure
Check RS232 baud rate or IEEE protocol selection for RS232
Incorrect command set selected
Check correct command set is selected
No head movement in manual mode
Power loss
See above
No output signal received by the CMM computer
Check connections and integrity of cabling from:
- head to PHC10-2
- PHC10-2 to interface
- interface to CMM computer
Multiwire bypass connector not fitted
When using standard touch-trigger probes, the probe head multiwire bypass connector must be fitted to permit the probe signal to reach the PHC10-2
Multiwired probe in use
Check that the multiwire cable is correctly fitted to the head
Check that the trigger output to the CMM computer is connected to the multiwired probe interface
Poor measurement performance
Loose mounting of head
Ensure all mounting screws are tight and mounting to CMM is secure
Probe DAMPing enabled during measurement
Ensure probe DAMPing is not enabled during measurement moves