Air supply
It is important that the air supply to the head is maintained while the head is powered as the air supplies the bearings and provides air for thermal management.
REVO-2 and air filter - labelled
CAUTION: Air supply pressure at the input to the air filters should be a maximum of 8.5 bar and at the head should be a maximum of 5.5 bar.
Air filter maintenance
WARNING: It is vital that the supplied air filters are used for REVO-2 and that the filter unit is maintained to the filter manufacturer's specification. Failure to comply with the specification will invalidate the warranty and is likely to result in permanent damage to the REVO-2 head requiring replacement at full cost.
To maintain the air quality:
1. Ensure the air provided to the filtration system meets the specification given.
Incoming air supply specification:
Air pressure
5 bar - 8.5 bar*
Air consumption
Up to 50 litres per minute
BS ISO 8537-1:2010 [4:4:4]
Particle size
15 μm
Dirt concentration
8 mg / m³
Dew point
+3 °C
Oil
5 mg / m³
* NOTE: The exact values are dependent on the inherent characteristics of the air pipe between the air filters and the REVO-2 head. Please ask the installer for the specification.
If the air pressure falls outside the range given, the head metrology results will be affected.
2. Regularly monitor the dew point indicator (see image below) and follow the instructions given in the table.
Maintenance of the REVO-2 air filters
For more information please visit the air filtration manufacturer's website http://www.smcpneumatics.com and search for IDG5 filters. (Please refer to the section 'REVO air filter service kits' for details on air filter service kit part numbers).
Dew point indications and maintenance instructions
Grain colour
Causes
Solutions
Blue / green
Normal operation.
No action required.
Pink / yellow
Water and oil flow into membrane air dryer.
1) Check and replace filters if necessary.
2) Check condition of inlet air for excess oil or water.
NOTE: The grains may be pink / yellow on delivery and can take up to 1½ hours to turn blue / green when air is connected.
NOTE: When dew point indicator absorbs vapourised oil content or other gaseous components in the compressed air, it may turn a colour other than blue / green or pink / yellow.
Air filter service kits
The Renishaw filter service kit (M-3060-0933) contains:
SMC part number
Description
Quantity
AF20P-060S
5 μm filter element for use with AF20 unit
1
AFM20P-060AS
0.3 μm filter element for use with AFM20 unit
1
AFD20P-060AS
0.01 μm filter element for use with AFD20 unit
1
We recommend that the above filter kit is replaced every two years.
The Renishaw membrane dryer service kit (M-3060-0943) contains:
SMC part number
Description
Quantity
IDG-EL5
Membrane module kit for use with IDG5 unit
1
IDG-DP01
Dew point indicator kit for use with IDG5 unit
1
We recommend that the above filter kit is replaced every four years.