SPA3 troubleshooting
Fatal faults
The SPA3 monitors internal operational functions. If any of these reach a defined level where the system believes it is dangerous to operate a fatal fault will be applied.
When a fatal fault is detected it is not possible to engage the servo system. There may be other faults that prevent the CMM from engaging.
Reported fatal faultPossible cause
Emergency stop
Emergency stop system activated
Internal failure of servo power amplifier
Air pressure
Air pressure low
Crash
Crash switch activated
Scale failure
Scale readhead failed
Scale “mis-count”
Scale not connected
Overspeed
The position demand has exceeded an internal defined limit
Failure of a tacho feedback
Incorrect scale resolution
Over temperature shutdown
The SPA3 unit is designed for continuous power of up to 800 W. The unit will allow a supply current consumption of up to 12 A, which at 80 V is 960 W for a short period of time (1 second).
If this is sustained for a prolonged period, or the unit is not used in an adequate environment (such as one with poor ventilation or high ambient temperature) the unit will disengage the motors to protect itself. An over temperature warning will be reported to UCCserver.
SPA3 visual diagnostics
Legacy SPA3 front panel
A visual indication of the system status is provided by three LEDs on the front panel, providing assistance in diagnosing and rectifying system faults.
LED status key
Description
SPA3-2 LED visual diagnostics - green yellow red
LED on
SPA3-2 visual diagnostics - Red flash
Red flash
Green flash
SPA3-2 LED visual diagnostics - green and red
Alternating green and red flash
SPA3-2 visual diagnostics - no LED
LED off
Any condition
Servo statusProbe seatedErrorDescription
SPA3-2 visual diagnostics - no LEDSPA3-2 visual diagnostics - no LEDSPA3-2 visual diagnostics - no LED
No power to SPA3
SPA3-2 LED visual diagnostics - redSPA3-2 visual diagnostics - no LEDSPA3-2 visual diagnostics - no LED
UCC2 connected, waiting for download, or not connected to a controller
SPA3-2 LED visual diagnostics - redSPA3-2 LED visual diagnosticsSPA3-2 visual diagnostics - no LED
E-STOP active, or power amp has a fault
SPA3-2 LED visual diagnostics - redSPA3-2 LED visual diagnostics - greenSPA3-2 visual diagnostics - Red flash
UCC connected, waiting for download, or download failed
SPA3-2 LED visual diagnostics - yellowSPA3-2 LED visual diagnosticsSPA3-2 visual diagnostics - no LED
Amps not engaged and no amp fault
SPA3-2 LED visual diagnostics - greenSPA3-2 LED visual diagnosticsSPA3-2 visual diagnostics - no LED
Amps engaged
SPA3-2 LED visual diagnostics - green and redSPA3-2 LED visual diagnosticsSPA3-2 visual diagnostics - no LED
Dynamic braking