Retrofit cabinet installation
The installation of the retrofit cabinet should only be performed by Renishaw or a Renishaw certified installation engineer.
Connection of CMM to the cabinet is either through an interface PCB or directly to the UCC / SPA units inside.
Refer to the relevant installation guides for the equipment to be fitted:
www.renishaw.com/cmmguides
To ease installation it is possible to remove the side panels from the cabinet by unlocking them with the keys provided.
The cabinet can be used in conjunction with the Renishaw retrofit desk.
CAUTION: The control cabinet, whether empty or equipped, exceeds the limits for a two person lift. Please follow the instructions on the outer packaging in order to remove the cabinet from the pallet. The cabinet has been provided with integral wheels for moving it around. It is not permitted to lift a cabinet with lifting equipment - no lifting points are provided.
The installation engineer is responsible for positioning the cabinet correctly. Should it need moving you must do the following:
  • Ensure a minimum clearance of 300 mm around all sides
  • Keep cables tidy, secure from being easily pulled out and not anywhere where they may present a trip hazard
  • Keep the working environment within specified environmental conditions
  • Always match the cabinet to the CMM as the cabinet is configured specifically for each CMM installation
  • Lock the cabinet wheels to prevent accidental movement
  • Review the risk assessment
  • A unique set of keys are provided with the retrofit cabinet (spares are available from Renishaw)
Electrical supply ratings
Retrofit cabinet model typeVoltageFrequencyCurrentTransient overvoltage categoryFuse type
A-5785-0100230 Vac ±10%50 / 60 Hz10 A maximum*210 A quick blow fuse HBC 250 V (5 mm × 20 mm)
A-5567-1000110 Vac / 230 Vac ±10%50 / 60 Hz10 A maximum*210 A quick blow fuse HBC 250 V (5 mm × 20 mm)
A-5785-0200110 Vac ±10%50 / 60 Hz16 A maximum*216 A quick blow fuse HBC 250 V (5 mm × 20 mm)
* It is the responsibility of the equipment installer to ensure that this maximum is not exceeded.
Refer to the label located near the mains input connector.
WARNING: Only use the installed mains supply cords.
Mains service fuse replacement
NOTE: A blown fuse indicates a defect within the cabinet. Please contact the installer or your local Renishaw office for advice.
C14 input connection
Retrofit cabinet - fuse locations - C14
CEE industrial socket input
Retrofit cabinet - fuse locations - CEE
Protective bonding (earthing)
WARNING: The site operator is responsible for providing adequate earthing provisions.
CAUTION: Any interruption of the protective conductor can make the equipment dangerous.
This equipment must be connected to a protective earth conductor via a three core mains cable. The mains plug should only be inserted into a socket outlet with a protective earth contact.
The CMM should be bonded to the cabinet to provide an equipotential earth connection.
Make sure that the grounding requirements are strictly observed.
Mains input connector
The chord used for connection to the mains must be rated to the minimum of supply specification detailed in 'Specifications'.
