Machine I/O
The 44-way (high density) socket provides connection for the CMM I/O.
The table below details the I/O pinout:
Pin no.FunctionDescription
1
+24 Vdc
Supply provided for use on CMM switched
2
Air
Output to control an air solenoid
3
X brake
Output to control the X brake
4
Y brake
Output to control the Y brake
5
Z brake
Output to control the Z brake
6
Output 0
General purpose output signal
7 to 12
Reserved
13
Input 0
General purpose input signal
14
Input 1
15 to 19
20
De-clutch
CMM de-clutch signal
21
Reserved
22
Low air
Low air pressure switch input
23
Crash
Crash detector switch input
24
Reserved
25
X axis outer limit
CMM X axis limit switch input
26
- X axis outer limit
27
Y axis outer limit
CMM Y axis limit switch input
28
-Y axis outer limit
29
Z axis outer limit
CMM Z axis limit switch input
30
-Z axis outer limit
31
X axis inner limit
32
-X axis inner limit
33
Y axis inner limit
34
-Y axis inner limit
35
Z axis inner limit
36
-Z axis inner limit
37
24 V return (0 V)
Common reference line for input and output signals
38 to 43
Reserved
44
24 V return (0V)
Shell
Screen
+24 Vdc
+24 Vdc limited to 1 A.
Air
Pin 2 will be be set low (to energise the air solenoid) on boot-up and subsequently, provided there is no air failure. They will be set high on this transistor output signal active low controlled by the “engage command”.
Brakes
During the commissioning process it is possible to invert the output from the brake connections from an active low signal to an active high.
General-purpose output
This is an open collector pulled down to the 0 V (24 V return) line via an NPN transistor when activated. It is suitable for driving devices in the range 5 V to 24 V with a maximum output current of 200 mA. If the output is not required, then it should be left open circuit. The output is protected for excess voltage and over-current within the controller.
Switched inputs
Two uncommitted inputs are provided. Each input is activated by being pulled down to the 0 V (24 V return line). The inputs are not opto-isolated and are pulled up to 24 V via a 10 kohm resistor. These inputs can accept signal levels in the range of +5 V to +48 V, and need to be pulled below 1 V to signify a change in state.
De-clutch
This input on pin 20 should normally be pulled down to the 24 V return line, to signal to the controller that the CMM's drives are mechanically connected to the moving elements of the machine, i.e. in their normal condition. The input should be open circuit, or held high, when a CMM has been temporarily ‘de-clutched' from its drive motors to allow manual positioning. The servo system will not drive the motors in this condition and will resume control at the machine‘s position when the signal again goes low.
Low air pressure
The low air pressure signal should be connected to a suitable air pressure switch. This input is monitored by the controller and when activated will remove power from the motors by causing a system fatal fault. During the commissioning process it is possible to invert this signal from an active low signal to an active high. If this capability is not required for integration to the system, then it should be connected to the 0 V (24 V return line).
Crash
The crash signal is used to inform the UCC T3 of a collision of the mechanical structure of the machine. Typically this is a sensor that is fitted to the end of the CMM quill and if activated will remove power from the motors by causing a system fatal fault.
Limit switches
The UCC T3 controller supports both inner and outer limit switches for all directions of machine axes. During the commissioning process it is possible to invert this signal from an active low signal to an active high. If this capability is not required for integration to the system, then it should be connected to the 0 V (24 V return line).
Inner limit switches
If an inner limit is activated, the UCC T3 system will perform a maximum movement deceleration in all axes to a stop and then perform a controlled axis back-off in the opposite direction as if a trigger event occurred.
Outer limit switches
If an outer limit is activated the UCC T3 system will immediately disengage all servo power to the machine drives. Recovery of this error is only possible by removing the outer limit switch activation, i.e. by moving the machine off the switch.
0 V (24 V return)
This is the 0 V reference for all the I / O signals.