Skip navigation

Installation procedure

WARNING: Ensure the PI 200-3 is disconnected from the power adapter during installation.

WARNING: Take care not to exceed the operation ambient of 50 °C around the unit. Do not install near sources of heat. Forced cooling may be required in final installation.

Inserting PI 200-3 card into PHC10-3 PLUS

Remove the end panel of the PHC10-3 PLUS with a hex-key and slide out the blanking plate.

Insert the PI 200-3 card, ensuring that the card is located in the correct grooves shown in the illustration below.

Insert the reset actuator over the top of the reset button on the PI 200-3 card and then replace the end panel.

PI 200 card kit contents
Removing PHC10-3 end panel
Inserting PI 200 card in correct location
Inserting PI 200-3 card reset actuator
Ensuring PI 200 card is secured properly
Replacing PHC10-3 end panel

Upgrading to TP200 (retrofit)

To successfully retrofit a TP200 probe system, the following points will need to be considered:

  • The CMM controller must service the PICS - PDAMP control signal - upgrade of the controller may be necessary if this signal is not available
  • The probe signal wires from the M8 connector on the probe head must be routed directly to the PI 200 interface - if an in-line interface for kinematic switching probes (TP1 / TP2 / TP6) is present, it must be removed
  • The signal wire resistance must be less than 5 ohms per conductor
  • Some experimentation with typical stylus arrangements may be necessary to obtain a satisfactory setting for the PDAMP / HALT filter delay time - the trigger / reseat debounce time and other configuration options will also need to be set (refer to the‘configuration switches' section for more information)
  • If upgrading from TP2, measurement programs may need to be changed to suit the length of the TP200, which is 5 mm longer than the TP2
  • If upgrading from TP2, star or offset (cranked) stylus arrangements will require a minimum 5 mm extension piece to be added for use with the TP200
  • If cables other than those supplied by Renishaw are used, equivalent cable screening and connector bonding must be used to maintain compliance with the EMC standards

System interconnection

Interconnection diagrams for the most common product combinations. Consult the appropriate product installation guide for information about setting the configuration switches of the PHC10-3 PLUS or ACC2-2 system controllers:

  • PH10 PLUS series installation and user's guide (Renishaw part number H-1000-7592)
  • Autochange system installation guide (Renishaw part number H-1000-6010)

PI 200-3 interconnection with MH8

PI 200-3 and MH8

PI 200-3 interconnection with MIH head

PI 200-3 with MIH connection diagram

PHC10-2 and PI 200-3 interconnection with PH10M

PHC10-2 and PI 200-3 interconnection with PH10M PLUS

PI 200-3 card with PHC10-3 PLUS interconnection with PH10T PLUS and SCR200

PI 200-3 card with PHC10-3 PLUS interconnection with PH10T PLUS and SCR200

PI 200-3 with PHC10-3 PLUS interconnection with PH10M PLUS and SCR200

PI 200-3 with PHC10-3 interconnection with PH10M PLUS and SCR200

PI 200-3 card and PHC10-3 PLUS interconnection with PH10 PLUS and AC3

PI 200-3 card and PHC10-3 PLUS interconnection with PH10 PLUS and AC3

PI 200-3 and PHC10-3 PLUS interconnection with PH10 PLUS and AC3

PI 200-3 and PHC10-3 PLUS interconnection with PH10 PLUS and AC3

Using TP200 with the ACR1 autochange system

The CMM controller must assert the probe damping signal (PDAMP) on the PICS port when the probe is approaching the rack, to prevent a false trigger being generated when the probe contacts the docking port lid. If this is not possible, a shock absorbing pad (part number A-1085-0294 supplied with ACR1) must be fitted to the autojoint connector on the probe head at the point of contact with the port lid.

Shock absorbing pad fitted to PH10M

SCR200 stylus change rack

PI 200-3 and two SCR200 racks

The miniature DIN connector on the base of the SCR200 rack is connected to the SCR200 connector on the PI 200-3 rear panel using cable PL63 / PL64 / PL65 according to the length required. For applications requiring two racks, a dual rack splitter cable is needed. SCR200 adaptor cable must be fitted at the PI 200-3 end.

PI 200-3 and SCR200

Interconnection cables


Cable number

Cable ident

Length

Part number

Notes

1

PL1

260 mm to 710 mm (10.24 in to 27.95 in)

A-1016-0004

Coiled

1

PL2

410 mm to 1.27 m (16.14 in to 50 in)

A-1016-0006

Coiled

1

PL3

680 mm to 2.32 m (26.77 in to 91.34 in)

A-1016-0012

Coiled

1

PL4

4.5 m (177.17 in)

A-1016-0001

Plain

1

PL27

225 mm to 450 mm (8.86 in to 17.72 in)

A-1016-6370

Coiled

1

PL29

740 mm to 1.48 m (29.13 in to 58.27 in)

A-1016-6420

Coiled

2

PL22

0.4

A-1057-0131

3

PL5

400 mm to 800 mm (15.75 in to 31.5 in)

A-1016-0131

Coiled

3

PL6

800 mm to 1.6 m (61.5 in to 62.99 in)

A-1016-7673

Coiled

3

PL12

100 mm (3.94 in)

A-1016-7674

Plain

3

PL13

100 mm to 200 mm (3.94 in to 7.87 in)

A-1016-7675

Coiled

4

PLM6

6 m (236.22 in)

A-1016-7564

Unterminated on end

4

PLM7

4 m (157.48 in)

A-1016-7563

Unterminated on end

4

PLM8

6 m (236.22 in)

A-1016-7677

4

PLM9

4 m (157.48 in)

A-1016-7678

5

PL25

300 mm (11.81 in)

A-1016-0124

6

PL63

5 m (196.85 in)

A-1016-7630

6

PL64

10 m (393.7 in)

A-1016-7631

6

PL65

20

A-1016-7632

7

PL38

25 m (984.25 in)

A-1016-7625

Unterminated on end

7

PL42

15 m (590.55 in)

A-1016-7624

Unterminated on end

7

PL56

12 m (472.44 in)

A-1016-7626

Unterminated on end

7

PL44

8 m (314.96 in)

A-1016-7627

Unterminated on end

7

PL46

3.7 m (145.67 in)

A-1016-7628

Unterminated on end

7

PL45

1.8

A-1016-7629

Unterminated on end

8

PL70

2

A-1016-7634

9

PL76

1

A-1016-7643

10

PL112

1

A-2172-0004

11

PL97

0.25

A-1016-7660

12

PL63

5 m (196.85 in)

A-1016-7630

12

PL64

10 m (393.7 in)

A-1016-7631

13

PL65

20

A-1016-7632

14

PL172

0.75

A-5684-1000


Rack mounting

Mounting PI 200-3

PI200-3 front on mounting

To mount the PI 200-3 in a 19 inch cabinet, two 1/3 blanking panel kits are required (Renishaw part number A-1018-0179). Fit the panels using the M5 × 6 mm long screws provided.

Mounting PI 200-3 with PHC10-2

PI 200 with mounting to PHC10-2

To mount a PI 200-3 next to a PHC10-2 controller in a 19 inch cabinet, a bracket kit is required (Renishaw part number A-1018-0173). Fit the panels using the M5 × 6 mm long screws provided.