Appendix 1 - Probe signals
It is strongly recommended that the Renishaw AC2 card is used to interface the SP600M probe. However, should you wish to attempt to power the probe and deal with its outputs by your own means, the following information is necessary.
Grounding arrangement
The SP600M has two 0 V connections, 0 V-power and 0 V-reference. It is recommended that they are connected to a star point in the interface and kept separate from the probe screen as shown below. The 0 V-power is used to power the internal circuitry and transducers. The 0 V-reference is a signal from which the outputs are referenced.
sp600_grounding
Screening
It is recommended that the screening arrangements are based on the Faraday cage principle, i.e. the screen is continuous throughout the system. The SP600M connector shell is connected to the body of the probe.
Power supply filtering
The power supply to the probe should be filtered as close to the interface output connector as possible to reduce the output noise to <20 mV RMS in the frequency range 0-7 KHz.
Measurement signal filtering
The input bandwidth is 700 Hz. The measurement signals should be filtered as close to the interface connector as possible to stop high frequency noise getting into the interface. It is recommended that a first order filter with a 3 dB cut-off frequency of ~700 Hz is used.
Cable terminations
The probe cable terminates at the high-density 'D' plug and carries the following signals:
HDD pin no.
Wire colour
Signal description
Signal type
11
Orange
X
Analogue signal
4
Violet
Y
Analogue signal
5
Yellow
Z
Analogue signal
6
Red
+ 12 V
Power output
7
Blue
- 12 V
Power output
12
Blue/White
0 V-reference
Power output
1
Brown
+5 V
Power output
3
Green
0 V-power
Power output
10
Coax inner
O/T 1
Signal output (not used)
8
White
Probe identification
Output
14
Black/White
Head LED anode
15
Brown/White
Head LED cathode
Shell
Screen
Grounding
Total number of connections = 13