HCU1 error reporting
The HCU1 LCD display gives information on head and system errors.
Datum error
Fault
The motorised probe head has not locked up to the required level of accuracy.
Action
Refer to your motorised head user's guide for further information concerning this error condition.
Obstruct error
Fault
The motorised probe head has been obstructed while rotating to the required position.
Action
Overload error
Fault
The motorised probe head head has been overloaded during a machine quill move.
Action
Refer to your motorised head user's guide for further information concerning this error.
Initialise head (PH50 only)
Fault
The PH50 probe head has been switched off whilst unlocked and the internal reference system has been lost.
Action
Head must be initialised.
See Home (PH50 only) page for instructions.
Head not present
Fault The motorised head is not electrically connected to the probe head controller.
Action Check the head is electrically connected.
Check cable connections.
Check cable integrity.
Contact your CMM supplier for further assistance.
HCU1 not recognised
Fault
There is a communications error between the HCU1 and the motorised head controller.
Action
Contact your Renishaw representative for further assistance.
No response from controller
Fault There is a communications error between the HCU1 and the probe head controller.
ActionCheck the cable connections.
Return both units to your Renishaw representative for service.
System error
Fault
Internal motorised head system failure.
Action
No user recovery action.
Return motorised head to your Renishaw representative for service.
FaultNothing visible on LED display
ActionAdjust the LED contrast adjustment wheel