RTP20 fault finding
RTP20
CAUTION:
RTP20 is designed for automated operation under direct control of the CMM software. To avoid possible collision during use, RTP20 must not be unlocked, re-orientated or locked by hand during normal operation.
Poor measuring performance
Possible causes
Remedy
Loose mounting
Ensure shank mounting screws are tight and mounting to CMM is secure.
Problem with TP20 probe module
Diagnosis of the probe unit should be carried out with axes correctly locked an without head adjustment between probe points.
Problem with swivelling / indexing unit
Diagnosis of the indexing unit should only be carried out following satisfactory probe performance.
No probe signal and / or no probe status LED
Possible causes
Remedy
Cable faulty / not connected
Check continuity of cabling from head to interface / machine control.
Probe interface faulty / not connected
Ensure correct connection of interface / machine control.
Poor repeatability (RTP20 only)
Possible causes
Remedy
Axes lock procedure incorrect
Using automated cycle ensure lock lever is fully rotated to lock position.
Axes unlock procedure incorrect
Using automated cycle ensure lock lever is fully rotated into unlock position during indexing.
Forces imparted onto head during lock-up
Using automated cycle unlock and re-lock.
Attempted lockup in incorrect unqualified position
Using automated cycle unlock, reposition correctly and re-lock.
Axes ‘rattling' during indexing (RTP20 only)
Possible causes
Remedy
Incorrect unlock procedure
Using automated cycle ensure lock lever is fully rotated to unlock position.
Incorrect indexing procedure
Using automated cycle index each axis separately.
NOTE: The RTP20 is not user serviceable and should be returned to Renishaw if suspected faulty.
TP20
Possible causes
Remedy
Stylus configuration too long or not rigid
Use shorter stiffer stylus configuration.
Poor stylus assembly
Ensure stylus joints are kept to a minimum and the joints are clean and secure.
Contamination / damage to stylus ball
Inspect for damage, clean thoroughly with solvent.
Trigger force too high.
Use lower force module.
Unwanted triggering during probe or CMM movement
Possible causes
Remedy
Trigger force too low / stylus configuration to heavy
Use higher force module / reduce mass of stylus configuration.
Probe fails to rearm after trigger
Possible causes
Remedy
Trigger force too low / stylus configuration too heavy
Probe reseat failure
Re-trigger probe. If problem persists please return to Renishaw for service.
Loss of measuring accuracy
Possible causes
Remedy
Mounting not secure
Check that the RTP20 is correctly mounted on the shank and that the screws are secure.
Check clamping mechanism in the CMM quill is secure for RTP20.
RTP20 not fully locked
Ensure that the thumbwheel / lock lever is turned fully clockwise.
Force imparted to probe module after locking
Re-qualify the probe.
RTP20 worn or damaged
Use only with the specified probe and extension combinations.
Faulty probe module
Check by substitution or return to Renishaw or your supplier.